E1DA Cosmos ADCiso Review
Rate this ADC:
    Votes: 3 2.2%
    Votes: 1 0.7%
    Votes: 13 9.5%
    Votes: 120 87.6%
KSTR
Audio Precision is not selling an ADC. It is selling a measurement system with all of its features built-in. Members like to generate results that are comparable to mine. Since I use high performance sine analyzer, then they need to know if they can get there with this box as is.
The HPSA (High Performance Signal Analyser) with its outstanding specs is only used for single sine test signals as there is only one notch filter. So the question of the "pure" ADC performance is fully legit.
For multitone the normal ADC path is used, and that path is very comparable in performance to the stripped down signal path in the Cosmos ADC with the ES9822pro, in other words, the Cosmos ADC is on par with AP for anything else than pure sine.
pma
Suppression of aliases at 192kHz sampling frequency. Input signal is 186kHz sine from analog generator, 1.44Vrms. ADC input range is 2.7Vrms.
It is not "fully" legit. Many of the core tests are single frequency starting with FFT in the dashboard. THD+N vs level. THD+N vs frequency. THD+N vs power. And of course SINAD. Again, the goal is try to create comparable measurements to mine. For that you need to do well in these core tests.
pma
The key is in the autoranger, to compete in level sweeps. Distortion and noise at one selected range, at the rated level, would be similar, same or better than with the AP. Sweeps are not, for the noise reason.
However, ownership of the expensive test gear is not everything. And the ASR test set does not cover everything. To me, the goal is to find most revealing methods to examine the DUT to its limits.
KSTR
Which anti-aliasing filter did you use?
With "Linear Phase Apozing" I get perfect supression (down in the noise floor) of frequencies > fs/2 and and full scale input, at all sample rates up to 384kHz. Only very very close to fs/2 the alias needle is visible.
pma
I have made no filter selections. I installed the Cosmos driver according to its web page (I have win8.1 so the driver must be installed) and that was it. No filter selection.
Have you tried the input frequency close to Fs as I did?
welwynnick
In my early years I used play with lots of lab gear - oscilloscopes, sig gens, power meters and spectrum and network analysers.
I recall the latter had wide frequency and dynamic ranges, and I couldn't help wondering how well a modern signal analyser would fare as an audio analyser.
The vast majority don't go down below 9kHz; a few do reach 10 or 20Hz; but they're very expensive.
I hadn't realised, but there are actually quite a few dedicated audio analysers, with some familiar names, so I did a survey.
Typically, they have two analogue input/output channels that work over 10 to 80kHz, and many have options for digital I/O and additional channels.
There's a wide range of maximum voltage, and the I/O impedance is usually 100kohm/100 ohm.
The best ones I could find have a SINAD over 100dB, and they're all expensive (if they're even available at all).
The Rhode & Schwarz UPL models are also quite popular, but I couldn't find any specs (I guess it sits somewhere low down).
The Cosmos price includes the scaler so the input impedance is comparable, and the SINAD is Amir's stereo measurement (as they're all 2 channel).
As Audio Precision claims, nothing beats the APx555.
But I thought it would be useful to get some perspective on what E1DA and QuantAsylum had achieved, despite the best efforts of the big guns.
ManufacturerModelChannelsBandwidthAnalogue I/ODigital I/OPrice $Spec SINADTyp SINAD
Audio PrecisionAPx5552200kHzYY30,000117120
E1DACosmos2190kHzYN298117117
Audio PrecisionSYS2722280kHzYY26,000112115
Stanford Research SystemsSR12200kHzYY17,000109113
Agilent / HPU8903A2100kHzYO(6000)108110
Audio PrecisionAPx5252-480kHzYY27,000108110
KeysightU8903B2-880kHzYO28,000108110
QuantAsylumQA403280kHzYN600105110
Rhode & SchwarzUPV2-8250kHzYO12,000105110
Audio PrecisionAPx516280kHzYY6,000109109
PrismSound / SpectraldScope III2200kHzYY4,400105108
PrismSound / SpectraldScope M1290kHzYY4,400105108
Rhode & SchwarzUPP280kHzYY(7000)100105
QuantAsylumQA401280kHzYN450101101
staticV3
Audio PrecisionAPx555280kHzYY30,000117120
E1DACosmos2180kHzNN298117117
Agilent / HPU8903A2100kHzOO(6000)108110
KeysightU8903B2-880kHzOO28,000108110
QuantAsylumQA403280kHzNN600105110
Rhode & SchwarzUPV2-8250kHzOO12,000105110
QuantAsylumQA401280kHzNN450101101
Nice comparison table!
One slight correction is that with a FW update, the Cosmos supports up to 768kHz Fs. That expands the BW to 384kHz (flat FR only with compensation file attached below).
  • CosmosAdc_V13_768_phase_CFRcompensations.zip
KSTR
For the filter selection you need to download and run "Tweak_Cosmos.rar".
I now tried to replicate your plot, running 188kHz -6dBFS, @192kHz SR, 10++ averages and get basically the same result. Initially I had used short FFT size which did hide the alias a bit more.
A bit worse than spec (which says around -100dB or better for most filters), but still OK in my book.
Reactions: pma
Watire
Is -6dbFS a good test signal level for mainstream benchmarking ? Or reaching the closer to clipping level is the best practice ?
If a device plays better numbers at -12dBFS is it valid ?
Can you read the absolute voltage output level of a device with the Cosmos capture from the get go ?
GXAlan
You have to calibrate the Cosmos to actual voltage. 10V nominal was 10.2V for 0 dBFS for me.
As close to clipping without intersample overs is the best.
pma
One should not overrun the Cosmos with too sharp edges (dv/dt). Sine at 1Mhz near fullscale (which is admittedly brutal) gives
Yes. I did try that yesterday as well, but did not post. Those are the reasons why I posted another thread in 2022:
restorer-john
E1DACosmos2180kHzYN298117117
Where are you getting your bandwidth numbers from?
The APX555 analyzer tops out over 1MHz, and its analog gen goes out to over 200kHz.
The QA403 can do 384k (ADC only) for ~180kHz.
welwynnick
Yes the analyser input goes much higher, but I took the DAC output from here:
I think its fair to use the analogue output, so I changed the chart from 80 to 200kHz.
Bandwidth specs weren't easy to come by - often it's just sampling rate - but I doubt it's much of a discriminator for any audio analyser.
In each case, the optimum performance is only achieved over 20 - 20kHz.
KSTR
Actually relevant for music recording/playback is the range from -10dB to -30 or -40dBFS. You never run an ADC close to clipping in practice.
Tests at 0dBFS are not very meaningful with regard to real life situation.
However, DACs and ADCs that behave well near 0dBFS typically stay the same or get even better at lower levels. But not always, notably when there is a strong ESS-Hump present...
capslock
Maybe I've worked too much with AKM converters, but I have learned to stay away from close to clipping. ESS traditionally do better near fs than AKM devices if you use SINAD or THD+N as your metric. Looking at Amir's IMD vs. input level graph in the first post, I can see the curves flattening. So I suspect that as you get near fs, harmonics will increase but not as quickly as noise decreases.
For optimizing amplifiers, I/V-stages, preamp stages etc. I am always more interested in HD than in noise since I can average all I like. With ESS, though, I am never sure where the sweet spot is because of the hump. Pity it also occurs in ADCiso.
Concur. And as apples to apples goes, we should be comparing the AP in high perf mode to ADCiso with APU, which can be done at 1 and 10 kHz. My understanding is that the AP uses a gliding notch, which is something the APU clearly cannot do.
capslock
The other point that needs to be understood is an input impedance in the Cosmos mono mode when measuring a component with single-ended RCA output.
Though the impedance specified is 1kohm for 2.7V input range, it is valid only in case of balanced input of 1 channel in the stereo mode (R1 + R2 = 1kohm). In the mono mode, L and R inputs go parallel, reducing input balanced impedance to 500 ohm, but this goes again to half, 250 ohm, if we measure any single ended output device, because -In branch (R2, R4) is effectively shorted and we have R1//R3 as a result, thus 250 ohm only. Thus the voltage difference in my previous post. So, be very careful, comparing Cosmos ADC and APxxx distortion test results and level measurements when link level sources have been measured.
I know that Ivan @IVX declares a bit higher value of input impedance in the single-ended mode. I was curious and made measurements of the input impedance. In the single ended mode, 1 channel, I got 673 ohm for the left and 605.5 ohm for the right channel. In the single ended mono mode it makes 318.75 ohm, which is a bit more than 250 ohm estimated above. By simulating with OPA1632, I get 666 ohm in the single ended 1 channel mode and 333 ohm in the mono mode. This is pretty close to the measured values. I am curious if @IVX has anything to add ;).
SE mode:
I am surprised you got something that not only differed from the spec and simulation but was also different between channels. I suppose the input network uses 0.1% resistors just to obtain the CMRR. Even with 1%, I don't see how they could be that different.
restorer-john
Unequal source (output R) loading of the two legs to ground in SE/mono.
Top
