I want to be able to measure how high of a vrms my system is using, will the cosmos work for that?

Also, I notice voltage is adjustable on the back, what exactly does changing those do?
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
I want to be able to measure how high of a vrms my system is using, will the cosmos work for that?
You can set the Cosmos ADC to one of the eight input sensitivities using the DIP switches at the bottom of the unit, then go into REW's RTA window, click "calibrate level", enter the input sensitivity that the ADC is set to (e.g. "3.5"), and lastly set the unit in the top left corner of the RTA window to Volt.

That way, REW will tell you what the signal strength that's present at the ADC's inputs is in Vrms.

One problem: The Cosmos ADC's input sensitivities are not exact. For example, the 1.7Vrms input on my Cosmos ADC is more like 1.811Vrms.

Therefore, if you want precise Vrms readings, you should instead use a multimeter like the Aneng AN870 with a 50/60Hz 0dBFS test tone.
Its measurement error hovers around ±0.3-0.4% in the relevant voltage range:
chart_4-4x_foolhardy_Remacri.png
Additionally, you can use such a multimeter to calibrate REW so that all readings with the Cosmos ADC will display precise Vrms values.

Also, I notice voltage is adjustable on the back, what exactly does changing those do?
It changes the Vrms that has to be present on the Cosmos ADC's inputs to reach 0dBFS output signal strength. In other words: gain.
 
  • Like
Reactions: trl
OP
S

ShiZo

Addicted to Fun and Learning
Forum Donor
Joined
Sep 7, 2018
Messages
835
Likes
556
You can set the Cosmos ADC to one of the eight input sensitivities using the DIP switches at the bottom of the unit, then go into REW's RTA window, click "calibrate level", enter the input sensitivity that the ADC is set to (e.g. "3.5"), and lastly set the unit in the top left corner of the RTA window to Volt.

That way, REW will tell you what the signal strength that's present at the ADC's inputs is in Vrms.

One problem: The Cosmos ADC's input sensitivities are not exact. For example, the 1.7Vrms input on my Cosmos ADC is more like 1.811Vrms.

Therefore, if you want precise Vrms readings, you should instead use a multimeter like the Aneng AN870 with a 50/60Hz 0dBFS test tone.
Its measurement error hovers around ±0.3-0.4% in the relevant voltage range:
View attachment 210256
Additionally, you can use such a multimeter to calibrate REW so that all readings with the Cosmos ADC will display precise Vrms values.


It changes the Vrms that has to be present on the Cosmos ADC's inputs to reach 0dBFS output signal strength. In other words: gain.
Thank you. Is there a standard way to run the 0dbfs test tone?
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
Yeah, using a bit-perfect audio player to ensure that no normalization or limiting takes place.
Examples are foobar2000 with the Wasapi exclusive output mode on Windows, or UAPP/Hiby Music on Android
Or you can use the Tone Generator inside REW, in combination with a universal ASIO driver like FlexASIO.
 

Lambda

Major Contributor
Joined
Mar 22, 2020
Messages
1,793
Likes
1,532
Since you need a volt meter to calibrate the ADC in why not uses the volt Meter to messure the DAC output directly?
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
that's what I suggested.
Use a multimeter instead of the Cosmos for Vrms readings.

Bonus:
If you're buying a Cosmos anyway for measurements, then you can use the multimeter to calibrate REW for accurate readings with the Cosmos. That way you're not limited to 50/60Hz tones.
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
You will need a true RMS voltmeter to measure the input voltage at the ADC. You can’t measure only the output voltage at the DAC, because then you measure the open loop voltage of the DAC and not the voltage under load with the ADC. To calibrate the ADC to ADC sensitivity, for every input range you have to apply a sinus signal of 1kHz close to 0dB full-scale (Signal level that the ADC will not clip) and measure the voltage at the input of the ADC Just keep in mind the low input impedance of the ADC if you want to use it as a RMS Voltmeter. To measure higher voltages with the Cosmos ADC, you can use a voltage divider. Then you have to measure the input of the voltage divider (with the connected ADC) and give that voltage for calibration. A bit-perfect audio player isn’t required for calibration; you can use any 1 kHz signal source with low distortion. The accuracy of your voltmeter will finally give you the accuracy of your calibration. If a true RMS voltmeter is not available, the calibration could be also done with a 50 Hz signal and standard voltmeter, but lower accuracy.
The voltage at the ADC has to be measured between Pin 2 and Pin 3 of the XLR connector.
You may measure higher THD of your devices too because of the low input impedance of the Cosmos ADC.
 
Last edited:

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
@staticV3, can you please tell me any reason why a bit-perfect audio player should be used for a voltage calibration of an ADC from the side of an electronic engineer?
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
@staticV3, can you please tell me any reason why a bit-perfect audio player should be used for a voltage calibration of an ADC from the side of an electronic engineer?
There is none.

OP asked for a standard way of running the 0dBFS tone for measuring a DAC's output.
If you try to play that tone using Wasapi shared or some normal audio player for Android like foobar2000, then you will find that the DAC outputs significantly lower voltage than expected, due to the OS applying a limiter.

To make sure that you're measuring the actual voltage capabilities of your DUT, make sure to play that test tone using software which bypasses the default audio pipeline and sends the tone bit-by-bit.
 

Rja4000

Major Contributor
Forum Donor
Joined
May 31, 2019
Messages
2,773
Likes
4,731
Location
Liège, Belgium
The Cosmos ADC's input sensitivities are not exact. For example, the 1.7Vrms input on my Cosmos ADC is more like 1.811Vrms.
True.

On top of this, the Cosmos ADC has a quite low input impedance, decreasing with the range.

To measure a value, you try to limit the impact of your measurement on the value (there is always an impact, but you try to have it as little as possible).

That's why, typical Digital Multi Meter have input impedance in 1M Ohm range, and Amir's Audio Precision analyzer is around 200k Ohm.

The Cosmos ADC is rather in hundreds of Ohm.
 

abdo123

Master Contributor
Forum Donor
Joined
Nov 15, 2020
Messages
7,448
Likes
7,957
Location
Brussels, Belgium
You can set the Cosmos ADC to one of the eight input sensitivities using the DIP switches at the bottom of the unit, then go into REW's RTA window, click "calibrate level", enter the input sensitivity that the ADC is set to (e.g. "3.5"), and lastly set the unit in the top left corner of the RTA window to Volt.

That way, REW will tell you what the signal strength that's present at the ADC's inputs is in Vrms.

One problem: The Cosmos ADC's input sensitivities are not exact. For example, the 1.7Vrms input on my Cosmos ADC is more like 1.811Vrms.

Therefore, if you want precise Vrms readings, you should instead use a multimeter like the Aneng AN870 with a 50/60Hz 0dBFS test tone.
Its measurement error hovers around ±0.3-0.4% in the relevant voltage range:
View attachment 210256
Additionally, you can use such a multimeter to calibrate REW so that all readings with the Cosmos ADC will display precise Vrms values.


It changes the Vrms that has to be present on the Cosmos ADC's inputs to reach 0dBFS output signal strength. In other words: gain.
I don’t think any affordable voltmeter would be more accurate than 0.1 volts anyway tbh.

In general measuring voltage should be almost always comparative in audio. Never use absolute numbers and just compare A to B.
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
I don’t think any affordable voltmeter would be more accurate than 0.1 volts anyway tbh.
In post #16 of this thread, a customer compared voltage readings of his AN870 to a Keysight U1252:
Screenshot_20220602-232027_Chrome.jpg
Seems accurate enough for me tbh.

Edit: AC voltages are compared a few posts down
Screenshot_20220602-232524_Chrome.jpg
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
@staticV3, l dont agree, you can make a voltage calibration of the ADC with any low distortion sine source and not with a DAC only. When l measure the input voltage of the ADC l use HP 34401a with a basic accuracy of 0.06% AC. The resulution of REW for voltage indication is to slow with only two digits
 

staticV3

Master Contributor
Joined
Aug 29, 2019
Messages
8,058
Likes
12,935
?
I never claimed that a general purpose DAC was the one and only way of calibrating an ADC in REW, so not sure what it is that you're disagreeing with tbh
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
by the way, if your voltage calibration in REW is not correctly, the noise measuring of V/sqrtHz is incorrect too
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
Maybe we misunderstand, from my side of few the Cosmos ADC is for general purpose measuring in audio bandwith and not only for measuring DACs. As Ivan always says, Cosmos ADC will outperform AP 555x, but maybe only for one measurement. THD cant be measured correctly without the right now not available APU due to the too low input impedance of Cosmos ADC. Whats about measuring of levels different to 0.5 dB FS where Cosmos ADC is calibrated. Means measuring at other levels will give invalid measuring results. Ivan always tels THD compensation has to be corrected to level. So how to make correct THD vs level measuring with Cosmos ADC. Sorry. maybe again l am talking here at the wrong place in the forum
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
there is a temerature dependency of THD measuring with Cosmos ADC too and not all users have the thermostat board installed at their Cosmos ADC. I really dont know why the calibration is done at 27°C environmental temperature, because most users will not have that temperature in their room
 

G-K

Member
Joined
Jan 31, 2022
Messages
51
Likes
7
to be at the point to this part of the forum, as my own measuring results with cosmos adc, l couldnt see differences in level measuring results due to temperature
 

Lambda

Major Contributor
Joined
Mar 22, 2020
Messages
1,793
Likes
1,532
there is a temerature dependency of THD measuring with Cosmos ADC too and not all users have the thermostat board installed at their Cosmos ADC. I really dont know why the calibration is done at 27°C environmental temperature, because most users will not have that temperature in their room
it's way easier to heat things up then to cool them down...

Also defiantly the wrong thread for this start a new one if you have problems with your ADC.

I don’t think any affordable voltmeter would be more accurate than 0.1 volts anyway tbh.
You would be surprised how wrong you are abut this.
 