Measuring RME ADI-2 Pro with QuantAsylum QA401

Miska

Addicted to Fun and Learning
Audio Company
Joined
Feb 20, 2019
Messages
615
Likes
448
Anyone want to draw the input circuit :)

View attachment 23338

If you want clean audio band, you need to have >144 dB attenuation above Nyquist...

Looks like AKM ADC chip. Have you checked the data sheet how it's digital filter looks like at the sampling rate you are running at? The analog filter looks maybe 3rd order. And if you want fairly flat phase response in 100 kHz bandwidth it means fc around 400 kHz or so. Or if just for 20 kHz then fc around 100 kHz.

Then again, many ADC and DAC chips also exhibit noise modulation anyway... So could be that too.
 

March Audio

Master Contributor
Audio Company
Joined
Mar 1, 2016
Messages
6,378
Likes
9,319
Location
Albany Western Australia
Well ordinary audio interfaces seem to manage quite well without significant noise floor modulation so it's quite acheivable.

Matts information shows only a few dB change in noise floor which matches the AKM data, so something is amiss. I'm sure the digital filters are adequate. Whilst I'm interested in hearing what the ultimate problem is, I don't have time to persue it.
 
Last edited:

QAMatt

Member
Audio Company
Joined
Mar 1, 2019
Messages
29
Likes
100
If you want clean audio band, you need to have >144 dB attenuation above Nyquist...

Looks like AKM ADC chip. Have you checked the data sheet how it's digital filter looks like at the sampling rate you are running at? The analog filter looks maybe 3rd order. And if you want fairly flat phase response in 100 kHz bandwidth it means fc around 400 kHz or so. Or if just for 20 kHz then fc around 100 kHz.

Then again, many ADC and DAC chips also exhibit noise modulation anyway... So could be that too.

Hi Miska, it's a second order filter on the QA401--pretty much the same filter used by AKM in Figure 41 of the ADC spec and also on the AKM eval board.

The tradeoff here is CMRR. The more components you have in the path between in the input and ADC, the more you hurt CMRR, and CMRR is important to THD, especially in certain class D applications. So, yes, you could build a higher order anti-alias filter but you'll degrade other things that matter. There's no getting around that unless you have a lot of hand tuning on each board. But then you move to a much higher price tier, routine calibrations, etc.

And 105 dB of rejection is actually very workable. You can take a 300W class D amp, short the input, and run that amp's output straight into the QA401 inputs. Remember: even though no sound is playing, the amp's output is still slamming +/-50V before pumping through the output LC--and you'll measure very close to TI's published noise numbers for that amp (85uV A weighted, 20 to 20 KHz) in spite of TI using an AES17 filters.

A really quiet headphone amp might have noise at -110 dBV (unweighted, 20 KHz). To impact that measurement via aliased noise means your out of band noise would have to be nearly 1Vrms at a few hundred KHz. Or maybe 100 mVrms if you are looking across a hundred KHz bandwidth. If someone is building a product with that much noise at the higher frequencies (but dead quiet at audio frequencies), something is really, really wrong with the product--you could see the noise easily on a cheap oscilloscope.

PS. I think all (?) >20 bit audio ADC and DAC exhibit a rise in N+D as the output level increases near the upper limits. If they didn't, there would be no "valley" in the THD+N output curves. I think the particular issue here is whether or not the rise was excessive in the OPs QA401. In some DAC products, you can set your max output to be 10 dB below the 0 dBFS and avoid the valley (early RME ADI had 0 dBFS set at ~17 dBu and +4 dBu max I vaguely recall), but the core converter itself has the valley.
 

Miska

Addicted to Fun and Learning
Audio Company
Joined
Feb 20, 2019
Messages
615
Likes
448
OK, I'm a bit curious because those plots with QA401/AK5397 seem to have some similarities with ADI-2 that has AK5574 which seems to exhibit somewhat similar behavior what I've seen, but has 85 dB stop-band attenuation.
 

edechamps

Addicted to Fun and Learning
Forum Donor
Joined
Nov 21, 2018
Messages
910
Likes
3,620
Location
London, United Kingdom
I tried to reproduce the "rising noise floor" phenomenon today with my QA401 in loopback. Here's the difference between ~-28 dBFS and -~-1 dBFS:

qa401noise.png


If I understand correctly, what @March Audio is worried about is the difference in noise floor that I annotated with a red arrow.

Just for fun, I took the opportunity to design a custom THD+N vs. output level measurement procedure and I ran it against the QA401 in loopback:

qa401relative.png


The above is THD+N. If we only look at absolute N+D, then the "rising noise floor" is extremely obvious:

qa401absolute.png


We're still looking at around -98 dB THD+N when running near the top of the scale, though, which is good enough for many use cases.
 

amirm

Founder/Admin
Staff Member
CFO (Chief Fun Officer)
Joined
Feb 13, 2016
Messages
44,585
Likes
239,376
Location
Seattle Area
We really need to test the QA401 with a high performance analog generator rather than its own DAC. My AP has both options. Here is the difference:

1552853757938.png


As you see with the red curve, you get much better performance with the analo generator as levels get stronger. I hope to test the Q401 at some point which I get some breathing room.
 

March Audio

Master Contributor
Audio Company
Joined
Mar 1, 2016
Messages
6,378
Likes
9,319
Location
Albany Western Australia
I tried to reproduce the "rising noise floor" phenomenon today with my QA401 in loopback. Here's the difference between ~-28 dBFS and -~-1 dBFS:

View attachment 23778

If I understand correctly, what @March Audio is worried about is the difference in noise floor that I annotated with a red arrow.

Just for fun, I took the opportunity to design a custom THD+N vs. output level measurement procedure and I ran it against the QA401 in loopback:

View attachment 23781

The above is THD+N. If we only look at absolute N+D, then the "rising noise floor" is extremely obvious:

View attachment 23782

We're still looking at around -98 dB THD+N when running near the top of the scale, though, which is good enough for many use cases.

Thanks thats good information. My experience was that as you approached 0dB the noise floor (with a 32k FFT for example) would sit around -125dB with pretty much any DAC that I attached to it. By -60dB you will be down to below -150dB

Typical example here.

https://www.audiosciencereview.com/forum/index.php?threads/audio-measurement-gear.113/post-62413
 
Last edited:
D

Deleted member 2944

Guest
We really need to test the QA401 with a high performance analog generator rather than its own DAC. My AP has both options. Here is the difference:

View attachment 23784

As you see with the red curve, you get much better performance with the analo generator as levels get stronger. I hope to test the Q401 at some point which I get some breathing room.
I would be glad to loan you mine. I live close. :)

Dave.
 

Blumlein 88

Grand Contributor
Forum Donor
Joined
Feb 23, 2016
Messages
20,689
Likes
37,410
I tried to reproduce the "rising noise floor" phenomenon today with my QA401 in loopback. Here's the difference between ~-28 dBFS and -~-1 dBFS:

View attachment 23778

If I understand correctly, what @March Audio is worried about is the difference in noise floor that I annotated with a red arrow.

Just for fun, I took the opportunity to design a custom THD+N vs. output level measurement procedure and I ran it against the QA401 in loopback:

View attachment 23781

The above is THD+N. If we only look at absolute N+D, then the "rising noise floor" is extremely obvious:

View attachment 23782

We're still looking at around -98 dB THD+N when running near the top of the scale, though, which is good enough for many use cases.

Yes that noise floor change. For comparison here is a March DAC1 feeding the ADC in the Zen Tour from Antelope audio.
Blue is maximum level at 1 khz. Red is -18 db. Green is -60 db. You see distortion rise at max. You see some higher orders rise at -18 db which I wouldn't expect. But the ADC noise floor is not effected by the level changes in the signal.
March 1khz different levels ZT.png
 

QAMatt

Member
Audio Company
Joined
Mar 1, 2019
Messages
29
Likes
100
Yes, agree, edechamps, the N+D is the correct way to look at this. I've attached the N+D plots for both the AK5397 and AK4490. On the QA401, both ADC and DAC are 0 dBFS @ = +6 dBV, so if you connect in loopback, and then drive to 0 dBFS, then both are at the same point of operation, which would be at the very far right on the AK plots. In this region, according to AKM, there could be as much as a 29 dB rise in N+D. Some from harmonics, and some from noise. You'd see about 23 dB of that from the ADC, and 6 dB of that from the DAC.

But notice as you drop to -5 dBFS, the combined N+D degradation drops to just a few dB.

Blumlein 88, this behavior is in all sigma delta converters to some degree...the way you "hide" this in an audio DAC is you scale all incoming digital data by, say, -5 dB in the driver before you feed that to the hardware. In other words, if you send it a true 0 dBFS signal, the DAC driver would give it -5 dB of gain, and then it would output a -5 dBFS signal when you told it to output a 0 dBFS signal. And then you call that new lower signal 0 dBFS. And just like that, the DAC (or ADC) exhibits no growth to N+D at 0 dBFS. But in doing so, you throw away dynamic range/noise floor/etc....it's all a tradeoff. The QA401 does some of this (and thus avoid the full 29 dB penalty shown in the graph), but preserving the -115 dBV noise floor is important.

AmirM, I'm always happy to drop a unit in the mail on short notice when you get a time slot opened. Just let me know
 

Attachments

  • ak4490ak5397_NpD.png
    ak4490ak5397_NpD.png
    147.3 KB · Views: 197

proj964

Member
Joined
Mar 16, 2019
Messages
14
Likes
13
Location
MN
QAMatt said:
AmirM, I'm always happy to drop a unit in the mail on short notice when you get a time slot opened. Just let me know
amirm Hope you can take Matt up on his great offer!
 
Last edited:

amirm

Founder/Admin
Staff Member
CFO (Chief Fun Officer)
Joined
Feb 13, 2016
Messages
44,585
Likes
239,376
Location
Seattle Area
AmirM, I'm always happy to drop a unit in the mail on short notice when you get a time slot opened. Just let me know
Thanks. If you can spare one for 2 weeks or so, I can take it anytime. Hard to make a schedule for one as stuff keeps arriving and piling up. :)
 

amirm

Founder/Admin
Staff Member
CFO (Chief Fun Officer)
Joined
Feb 13, 2016
Messages
44,585
Likes
239,376
Location
Seattle Area
Its been over two months since QAMatt's offer. amirm did this ever arrive? Analyzed? ...
It arrived. I ran a few tests which I did not understand and ran it by Matt. Then the backlog of gear attacked me and I have not been able to get back to it. :) I will get to it soon.
 