Sure. They have independent parameters though (although share the gating setting)...
Same as CSD, I can change the parameters and get infinite variation of these graphs. You have to know what you are searching for before creating these graphs (e.g. deciding on time versus frequency resolution).
Okay, the possibilities are killing you
By default Arta, but also other measurement software scales to -30dB with 3dB or 1dB step per color. The color transitions should be as sharp as possible to increase readability.
Probably the best setting for the sonogram can only be found by trial and error.
Does such a diagram in the review really make sense?
It helps to classify irritations that appear in the Spinorama - is it a resonance, how severe is the resonance,...
Burst tones are supported but for a different purpose. There is no direct equivalent to this feature in ARTA.
I'm at a loss myself. Maybe the type of evaluation with the Klippel NFS has a different name.
Here is an extract from the Arta Manual:
No. There are two domains of analysis in Klippel:
1. Using their standard analyzer (KA3) and making simple measurements. Here, it is assumed you either use gated analysis or use anechoic chamber. Distortion, CSD, Sonograph, impedance, etc. all come out of this subsystem.
2. Robotic Near Field Scanner. This is a bolt on feature that uses #1 for measurements but then applies all the clever signal processing to model the soundfield and if asked, remove reflections.
There is a way to use #2 to remove reflections in measurements in #1. I have done a bit of testing on that and it indeed corrects the frequency response. I have not however looked to see if it also corrects the CSD, etc. to be reflection free.
With the approximately 30cm distance during the measurement, an evaluation up to 300-400Hz should be possible with gate.
If the Klippel NFS removes the room influences during frequency response, the evaluation of the CSD at lower frequencies should also be improved - I would assume.